Hello all.

 

We have a Ziess CMM. The probes are calibrated every morning, thus the device is unavailable for a good hour of production time. There is some disagreement as to whether daily calibration is necessary. What do you folks do? Calibrate daily or...

 

MCAM X6 MU1

Solids,Port Expert, Multi Axis, etc., etc.

Link to comment
Share on other sites

We do a re-qual on a fixed sphere and run a quick check program on a standard. The standard is a 2" and a 3" gage block set at angles in a fixture. If the machine measures the blocks within .0002 of nominal, we are good to go. The inspection report is filed and copies are included with part certification. The whole process, if everything measures out good, takes about 10 minutes.

Link to comment
Share on other sites

We calibrate all the probes at all orientations biweekly, but our cmm is in a climate controled room

that stays at 68 deg. The cal. takes about 2 hours to run. We will cal. more often if needed ie. dirty

probe or inconcistant results.

If you don't have a stable temp. or are probing dirty parts then you may want to do it more often,

but usually just a quick cal. on the first or primary probe in the morning will be ok.

 

HTH Ken,

Link to comment
Share on other sites

What requirements are you looking to meet? are you looking to simply verify the interim measurements then you could possibly do an interim artifact check at various times to prove validity of your measurements instead of a full calibration.

 

That brings you back to is the artifact checks equal to the measurements you are taking or needing?

 

What process and control are you going to use for the artifacts?

 

You could ideally run them as a sub program if conditions are ideal in your setups also with if then statements or scripting to generate the reporting you would like for your process.

Link to comment
Share on other sites

I calibrate the probe whenever I get readings I don't like. :harhar:

 

I've never run a Zeiss machine, but with PC-DMIS, I'll just calibrate the locations that we're going to use. Obviously, this won't work for you if you use a brazillion orientations every day - but for us, it's a 10 minute process or so.

Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

 Share

  • Recently Browsing   0 members

    • No registered users viewing this page.

Join us!

eMastercam - your online source for all things Mastercam.

Together, we are the strongest Mastercam community on the web with over 56,000 members, and our online store offers a wide selection of training materials for all applications and skill levels.

Follow us

×
×
  • Create New...